Heh. 400 MHz would be nice. I guess one of the compromises of having a 750 mm bore, is having a lower field strength/frequency - 133 MHz in this case.
Most big MRI scanners don't use LN2. The early ones did, but this was rapidly dropped, presumably for reasons of complexity. Helium reliquification equipment is standard on all new MRI scanners, which makes helium top-up an unusual maintenance activity. Without the reliquification system operational, boil off rates are typically between 200-400 ml/day, depending on field strength - higher fields tend to boil off faster.
1762 l is a lot of helium. I'm told that restarting the magnet after an emergency quench is expected to cost about $50k.