So today, real "balances" are far and few in between with one of the few exceptions being the beam balance used at doctors' office. They've been almost universally replaced by scales.
Scales don't use any reference masses, so they have to be frequently calibrated with masses. They also have issues with sensitivity being non-linear across the span.
Even with balances, there was an issue with differences in load affecting fulcrum friction, but someone came up with this awesome idea in mid 20th century that keeps the fulcrum at a constant load the whole time.
It looks something like this:
diagram
If the capacity is 200g, then there is 199grams of mass(made of a bunch of little masses) sitting above the weighing hanger.
If you're weighing 149.2522g, you put the beam on a heavy duty fulcrum, then turn the dials to remove about 149g of weight until it becomes about level.
At this point, you put the beam on the delicate fine weighing fulcrum, then turn the fine tuner (like moving the finger on a triple beam) until it becomes level, then you read it out. Add it to 149 and you get 149.2522g.
If you were measuring 49.2522g, then you remove the 100g weight from the top, but the fulcrum won't see the difference, since it's always under "full load".
These things never need calibration and the response is linear, as long as the "reference masses" are very accurate.
So, aren't these things better than an electronic one that tries to measure from 0.1mg to 150,000mg (1 to 1.5 million ratio) and needs constant calibration?
Scales don't use any reference masses, so they have to be frequently calibrated with masses. They also have issues with sensitivity being non-linear across the span.
Even with balances, there was an issue with differences in load affecting fulcrum friction, but someone came up with this awesome idea in mid 20th century that keeps the fulcrum at a constant load the whole time.
It looks something like this:
diagram
If the capacity is 200g, then there is 199grams of mass(made of a bunch of little masses) sitting above the weighing hanger.
If you're weighing 149.2522g, you put the beam on a heavy duty fulcrum, then turn the dials to remove about 149g of weight until it becomes about level.
At this point, you put the beam on the delicate fine weighing fulcrum, then turn the fine tuner (like moving the finger on a triple beam) until it becomes level, then you read it out. Add it to 149 and you get 149.2522g.
If you were measuring 49.2522g, then you remove the 100g weight from the top, but the fulcrum won't see the difference, since it's always under "full load".
These things never need calibration and the response is linear, as long as the "reference masses" are very accurate.
So, aren't these things better than an electronic one that tries to measure from 0.1mg to 150,000mg (1 to 1.5 million ratio) and needs constant calibration?